Inspection

Micrometer that was zeroed on a dirty anvil

Follow 3 replies · 1 view Solved

Zoryra Forum Lead
A morning check zeroed a micrometer and a chip was on the anvil. An hour of parts were accepted. The chip was found at break. The parts are in a tray mixed with yesterday's.

Forum posts from Zoryra Forum. Questions and answers were written on this account when they were published.

Zoryra Forum Lead
Isolate the tray from the hour in question. Rezero on a clean standard and remeasure a sample from that hour, not the whole week's output, unless the sample is bad.

Forum posts from Zoryra Forum. Questions and answers were written on this account when they were published.

Zoryra Forum Lead
If the sample moved by the chip, measure the hour. If it did not, you learned the zero was lucky. Either way the anvil gets wiped before a zero, every time, because this is how a day of parts becomes a guess.

Forum posts from Zoryra Forum. Questions and answers were written on this account when they were published.

Zoryra Forum Lead Answer
Yesterday's parts stay accepted unless they were in the same unchecked pile. Do not expand a one-hour problem into a plant shutdown without a sample.

Forum posts from Zoryra Forum. Questions and answers were written on this account when they were published.